This behavior is expected when using the protected version of the NI TB-2640B terminal block.
The TB-2640B is available in two hardware variants:
- P/N 782385-02 includes fixed 100 Ω protection resistors on each column.
- P/N 782385-01 does not include column protection resistors.
Both terminal block variants include 100 Ω protection resistors on the external row connections. These resistors can be bypassed using 2 mm jumpers installed on header J14.
On TB-2640B P/N 782385-02, 100 Ω resistors are permanently installed in series with each column to help protect the reed relays from high inrush currents caused by capacitive loads. These resistors cannot be bypassed using jumpers.
When two columns are connected to the same matrix row, the signal travels through the internal row bus rather than through the external row connections.
As a result:
- The signal path includes two 100 Ω column protection resistors connected in series.
- The measured path resistance is approximately 200 Ω.
- The J14 bypass header does not affect the measurement because the signal path does not pass through the external row connections.
To determine whether the behavior is expected:
- Check the part number printed on the TB-2640B label.
- Verify whether the installed terminal block is P/N 782385-02 or P/N 782385-01.
- If the terminal block is P/N 782385-02, approximately 200 Ω column-to-column path resistance is expected behavior.
- If the terminal block is P/N 782385-01, the path resistance should typically be less than 2 Ω.
Select the Appropriate Terminal Block for the Application
- Use TB-2640B P/N 782385-01 when low path resistance is required.
- Use TB-2640B P/N 782385-02 when switching capacitive loads or when additional relay-contact protection is desired.