Slew rate vs Current Change Time on NHR 9300
Updated Oct 26, 2026
I noticed that the specifications for the NHR-9300 list both a Slew Rate and Current Change Time. What are the differences between these specs, and how fast can I expect my NHR-9300 to go from 125A to 0A when operating in the high-voltage range?
Slew Rate describes the speed of the current transition over the 10-90% of the selected current range. It is intended to characterize the slope of the transition within that range, and does not represent the total time required to change from one setpoint to another. The 10-90% transition should occur within 2 ms in the 600V operating mode, and within 3 ms in the 1200V operating mode.
Current Change Time specifies the total time required to make a change between two setpoints in the same operating mode and polarity. Any setpoint change that does not cross 0A should occur within 5 ms.
For the example of changing the current setpoint from 125 to 0 A when in the higher-voltage mode, the Current Change Time spec would apply and you could expect this to occur in less than 5 ms. This assumes that you have not programmed a slower Slew Rate in the 9300 operating settings.
The response below 10% of the current range is not separately specified, so the exact behavior near 0 A is not guaranteed by the published specifications.
Waveform shape and settling characteristics can vary depending on programmed slew rate, operating mode, cabling, and sense-lead configuration.