The published DMM reading-rate specifications typically represent the maximum attainable measurement throughput under specific operating conditions and do not always reflect the timing required for resistance measurements.
For resistance measurements, the total measurement cycle includes multiple timing components:
- Internal switch time
- Settling time
- Aperture time
The NI-DMM measurement cycle defines settling time as the period required for the measurement circuitry and applied excitation current to stabilize before a valid measurement can be taken. Aperture time is the period during which the analog-to-digital converter (ADC) acquires the measurement. Both values directly contribute to the achievable measurement rate.
To estimate the maximum resistance measurement rate:
- Determine the configured resistance range.
- Identify the applicable settling time.
- Identify the configured aperture time or resolution setting.
- Calculate the approximate measurement rate as:
Measurement Rate ≈ 1 / (Settling Time + Aperture Time + Internal Switch Time)
For the NI PXI-4065, default resistance settling times vary significantly by range. Examples include:
- Resistance ≤1 kΩ: 10 ms
- Resistance 10 kΩ: 40 ms
- Resistance 100 kΩ: 200 ms
- Resistance 1 MΩ: 200 ms
- Resistance ≥10 MΩ: 750 ms
As resistance ranges increase, settling time becomes the dominant contributor to overall measurement duration. Consequently, resistance throughput may be substantially lower than the headline readings-per-second specification.
Aperture time also affects throughput. Longer aperture times improve noise rejection and measurement resolution but reduce measurement speed. NI documentation notes that shorter apertures generally increase reading rates, whereas longer apertures reduce throughput.
When comparing the PXI-4065, PXI-4070 family, and PXIe-408x family:
- Higher published reading-rate specifications indicate the maximum conversion capability of the DMM.
- Actual resistance measurement performance depends on the combination of settling time, aperture time, range selection, resolution, auto-ranging behavior, and application configuration.
- Resistance measurements on higher-resistance ranges may be limited primarily by settling requirements rather than by the instrument's advertised reading rate.
To maximize resistance measurement throughput:
- Select a fixed range instead of using Auto Range when possible.
- Reduce aperture time or resolution requirements where measurement accuracy permits.
- Minimize user-configured settling time while maintaining required measurement accuracy.
- Evaluate throughput using the complete measurement cycle rather than relying solely on the headline readings-per-second specification.