PXIe-5111 Voltage Measurement Clips Near 23 V with 30 V DC Input
Updated Oct 6, 2026
  • PXIe-5111
  • InstrumentStudio
  • NI-SCOPE
I am using an NI PXIe-5111 oscilloscope module to acquire a DC voltage signal from a Fluke 5550A Multi-Product Calibrator. When the calibrator output is set to +30.000 V DC, the waveform displayed by the PXIe-5111 does not reach the expected voltage level. Instead, the measured signal rises and plateaus at approximately +23 V.
I verified the calibrator output using a digital multimeter (DMM), which measures the expected +30.000 V DC. The behavior is consistently reproducible and occurs when the PXIe-5111 input impedance is configured for 1 MΩ. The same behavior is observed when both the calibrator and the PXIe-5111 are configured for a 1 MΩ impedance setting.
The NI PXIe-5111 has a maximum full-scale vertical input range of 40 Vpk-pk at a 1 MΩ input impedance. When the Vertical Offset is set to 0 V, the acquisition window is centered between -20 V and +20 V. A +30 V DC input exceeds this positive analog-to-digital converter (ADC) clipping limit and causes the measured waveform to saturate near +23 V.
To properly measure a +30 V DC signal, configure the channel settings in NI InstrumentStudio or the NI-SCOPE API:
  1. Open InstrumentStudio or your NI-SCOPE application.
  2. Set the channel Input Impedance to 1 MΩ.
  3. Set the Vertical Range to 40 V (or 40 Vpk-pk).
  4. Set the Vertical Offset to 20.0 V (or within +15 V to +20 V).
Adjusting the vertical offset shifts the measurable voltage window so that the entire +30 V DC signal falls within the 40 Vpk-pk dynamic range of the digitizer (for example, setting the offset to +20 V shifts the input span from 0 V to +40 V).