First, verify that the chassis in which the module is installed provides sufficient slot cooling capacity for the PXIe-5842.
According to the PXIe-5842 User Manual, the module requires a minimum slot cooling capacity of 82 W. Verify the chassis specifications and confirm that it supports this cooling requirement.
If the chassis meets the cooling requirements and the PXIe-5842 continues to report the nonfunctional state, try changing the chassis fan mode by using the chassis DIP switch settings. For the PXIe-1095:
- Power down the PXI system.
- Locate the DIP switch bank on the chassis.
- Identify the Fan Mode switch.
- Verify the setting:
- If the switch is set to High, change it to Auto.
- If the issue persists, test both Auto and High fan modes and verify whether the negotiated cooling value changes.
- Power the chassis back on and allow the system to boot.
- Open NI MAX and verify the PXIe-5842 status.
In some cases, changing the fan mode may resolve incorrect cooling negotiation behavior reported by the chassis.