Resolve RFmxSpecAn CHP Fetch Measurement Error -1074135025 on PXIe-5841
Updated Aug 28, 2026
  • PXIe-5841
  • RFmx SpecAn
  • NI-RFSA
I am evaluating a PXIe-5841 and using RFmx Spectrum Analyzer measurements together with NI-RFSA functionality in a custom session-management architecture.
After a successful measurement, repeated acquisitions may fail with the following error:
Error -1074135025 occurred at RFmxSpecAn CHP Fetch Measurement.vi
Possible reason(s):
IVI: (Hex 0xBFFA000F) Invalid parameter.
Parameter Name: startingRecord
Corresponding Value: 1
Maximum Value: 0
I may also observe invalid session or invalid reference errors when using RFmx and NI-RFSA together, particularly when performing resets between measurements. The first acquisition works correctly, but subsequent measurements fail or exhibit inconsistent behavior.
The observed behavior can occur when RFmx and NI-RFSA are sharing the same underlying analyzer session and the session state is modified by lower-level reset operations.
To resolve the issue:
  1. Initialize RFmx and NI-RFSA using a shared session architecture.
    • Use RFmxInstr Initialize NIRFSA.vi to establish the RFmx session.
    • Use RFmxInstr Get NIRFSA Session.vi when NI-RFSA access is required.
    • Maintain a single owner for the physical PXIe-5841 session whenever possible.
  2. Use a persistent device-level session manager.
    • Initialize the hardware once.
    • Keep the session active for the duration of operation.
    • Prevent multiple independent components from initializing, resetting, or closing the same device session.
  3. Avoid calling niRFSA Reset With Options within the normal measurement loop.
    • This operation can clear driver state expected by RFmx.
    • Repeated measurements may fail even when the session handle remains valid.
  4. Use RFmxSpecAn Reset to Default.vi when a software reset is required.
    • This approach preserves RFmx ownership of the underlying session.
    • Testing showed successful repeated execution after replacing the NI-RFSA reset operation with the RFmx reset method.
  5. Verify record-handling logic when performing repeated measurements.
    • For standard single-record acquisitions, fetch record 0 on each acquisition.
    • Do not advance startingRecord unless a multi-record acquisition has been explicitly configured.
  6. Reapply acquisition settings after any reset operation if repeated measurements are required.
    • Number of Records
    • Number of Samples
    • IQ Rate
    • Trigger configuration
    • Fetch-related settings
  7. Review the application for unintended NI-RFSA acquisition calls.
    • Check for direct IQ acquisition functions such as Fetch IQ, Read IQ, or other helper VIs that may be operating on the same session.
    • Ensure RFmx and NI-RFSA configurations remain synchronized.
Following these recommendations resolved the repeated measurement failures and allowed successful execution of RFmx Spectrum Analyzer measurements without the original error.
Technical observations from the investigation:
  • RFmx and NI-RFSA share the same underlying analyzer session.
  • Direct NI-RFSA reset operations can invalidate or modify state expected by RFmx measurements.
  • A centralized session-management model is recommended when multiple RF personalities or APIs access the same PXIe-5841 device.
  • Replacing niRFSA Reset With Options with RFmxSpecAn Reset to Default.vi resolved the reported issue during testing.
  • Additional validation may be required when operating the Signal Generator and Signal Analyzer simultaneously as independent instruments.