For example, if operating the PXIe-4311 in the -150 to 150mV range with a common mode voltage of 5V and a signal frequency of 10Hz, the common mode error expected would be:
Common mode error = CMRR*(common mode voltage)
Common mode error = (1.6uV/V)*(5V) = 8uV
Please note that the PXIe-4311's self-calibration procedure calibrates the module's internal path, but it does not remove system level common mode error. If a concern, the common mode error can be measured and compensated as described above.