The following example uses the offset specification for a calibrated NI-9232 within one year of calibration under typical conditions of 23 °C, ±5 °C. In the specification table, the offset value of 0.023% of the range is highlighted in red.
A separate range table shows that the selected typical input range is ±31.5 V peak.
Using a percent of range value of 0.023% and a typical range of 31.5 V, the offset voltage can be calculated as follows:
Offset Voltage = 31.5 V × (0.023 / 100)
Offset Voltage = 31.5 V × 0.00023
Offset Voltage = 0.007245 V
The expected offset voltage is 7.245 mV under these specification conditions.