Determining the Maximum Throughput of a CompactDAQ Chassis
Updated Feb 27, 2025
  • NI CompactDAQ Chassis
The maximum amount of throughput, or data transfer, for a particular CompactDAQ Chassis is both bus and system dependent. Below is a summary of the different types of CompactDAQ Chassis and how to calculate the maximum throughput for the hardware.
CompactDAQ USB Chassis
The NI X Series DAQ and CompactDAQ chassis with a USB 2.0 bus can sustain a throughput of 32 MB/s, while the CompactDAQ chassis with a USB 3.0 bus can sustain data transfer rates of 250 MB/s.
CompactDAQ Ethernet Chassis
CompactDAQ ethernet chassis have a streaming performance of 32 MB/s. Examples include the cDAQ-9181/83/85/87/89 ethernet chassis.
In order to determine whether or not a device can handle the throughput of a given application, you will need to calculate the amount of data that will be streamed across the bus during that application. An example for calculating the data transfer rate of a device is shown below.
For example, suppose you have a 14-slot cDAQ-9179 (USB 3.0) and want to determine whether or not it can sustain the throughput of running 14 NI-9234 modules, acquiring at the maximum sampling rate (51.2 KS/s), across all 4 channels of every device. You could calculate the total data that will be streamed across the 9179's USB bus using the equation below:
Calculation of total data totalling at 8.6MB/s
Because the total rate of data transfer during this application (8.6 MB/s) does not exceed the maximum throughput of the 9179's USB 3.0 bus (250 MB/s), there should not be bandwidth issues acquiring data at these rates on this device.
Note: The throughput calculation for Digital IO modules, such as the NI-9474, are same the as explained above, but for the Counter IO, such as the NI-9361, the data transferred from the CompactDAQ to the host PC over USB would depend on the type of measurement performed. For example when doing a edge counting task, the data returned will be a 32-bit Integer at the sample rate defined. Similarly, frequency data is a 4-byte integer and Pulse measurements are 8-byte values.
