Noise Floor Calculation For NI Sound and Vibration Device
Updated Jul 14, 2026
  • Sound and Vibration Device
  • NI-9234
  • PCI-4462
  • NI-9231
  • When we verified the noise level in the soundproof room, we found that the value measured using B&K microphones and NI-9234 was around 32 dB. There was no way to measure below 30 dB at a sampling rate of 51200 samples per second (S/s). However, when we used a third party sound level meter to test in the soundproof room, the actual result was around 24.5 dB. Why can't NI-9234 measure noise levels below 30 dB?
  • What is the noise level calculation for NI Sound and Vibration hardware, such as NI 9234?
Before we start the noise floor calculation, it is important to obtain below information for ease of calculation.
  • Sample rate of the signal
  • Sensitivity of the microphone (refer to the specifications of the microphone)
With the above information, we can go to the hardware specification, for example NI-9234 Specifications, with a sample rate around 51.2 kS/s, the idle channel noise is 50 μVrms.
For a microphone with sensitivity = 50 mV/Pa, the channel's acoustic noise level is calculated below from the idle-channel noise (NI 9234 idle-channel noise = 50 µV rms):
Acoustic Noise Level = Idle channel noise / Microphone Sensitivity
Acoustic Noise Level = (50 μV rms) / (50 mV/Pa) = 0.001 Pa rms
Acoustic Noise Level = 20 * log10(0.001 Pa rms / 20 μPa rms*) = 34.0 dB SPL
A-Weighted Noise Level = 34.0 dB SPL - 2.8** dB = 31.2 dBA
For the same microphone (S = 50 mV/Pa), we can also calculate the acoustic noise floor from the integrated noise density (NI 9234 noise density = 310 nV/√Hz at Fs = 51.2 kS/s):
Noise Power = Frequency Bandwidth * ( Spectral Noise Density )²
Noise Power = 25.6 kHz  (310 nV/√Hz)² = 2.460n V² rms
Noise Level = √(2.460n V² rms) = 49.6 µV rms
Acoustic Noise Level = (49.6 μV rms) / (50 mV/Pa) = 0.000992 Pa rms
Acoustic Noise Level = 20*log10(0.000992 Pa rms / 20 μPa rms*) = 33.9 dB SPL
A-Weighted Noise Level = 33.9 dB SPL - 2.8** dB = 31.1 dBA
Notes:
*  Reference sound pressure level (SPL) is 20 µPa rms.
** For a measurement bandwidth of 0 to 25.6 kHz, A-weighting reduces the SPL by approximately 2.8 dB.