Voltage Drop in PXIe-4309 During Multi-Channel Measurements for Strain Gauges
Updated Nov 4, 2025
  • PXIe-4309
Why does a voltage drop occur when measuring multiple channels for strain gauges with the PXIe-4309 as the sample rate increases, but not when measuring a single channel?
The PXIe-4309 is equipped with an anti-aliasing filter and a multiplexer (MUX) for channel groups.
Strain gauges need to charge the input capacitance of the amplifiers as the inputs switch. For example, refer to the source impedance errors page in the specification document. Considering that strain gauges are generally greater than 120 ohms, this is likely the cause. The PXIe-4309 is optimized/guaranteed to work at ≤ 50 ohms and ≤ 10 S/s.
The anti-aliasing filter reduces voltage when measuring high-frequency components. However, this phenomenon also occurs in single-channel measurements, so it is not the cause in this case.