How To Measure LTE TDD Signal By RFmx LTE/LTE-Advanced
Updated Jul 4, 2024
  • RFmx LTE/LTE-Advanced
If you used RFmx LTE/LTE-Advanced to measure one LTE TDD signal with default setting , you could not get the correct result ; How could you set the RFmx panel to measure the LTE TDD signal ?
Below is one example about how to create one LTE TDD signal by RFmx Waveform Creator :
The LTE TDD signal have DL/UL configuration as 2:7 , which means the subframes as DSUDD DSUDD; There are two subframes working on PUSCH, subframe1 is QPSK and subframe2 is 16QAM ;
Open RFmx LTE/LTE-Advanced to measure and demodulate the signal with default setting , RFmx could not demodulate it correctly , like below picture :
How could you get the correct measurement result ?
1.Setting the Trigger mode as Digital Edge (setting SOURCE as PXI_Trig0 which is same as RFSG setting)
   Setting the SYNCHRONIZATION MODE as Marker
2.To get correct result of ModAcc , you should set the MEASUREMENT OFFSET as 4 Slots for subframe1;
You can see the correct constellation result for subframe1 (QPSK).
Of course , when you set  set the MEASUREMENT OFFSET as 14 Slots for subframe2; You also see the expected result about constellation for subframe2 (16QAM);
3.To get the correct result of ACP, you should set the correct value of Sweep Time; Change the default value 1.000ms as 10ms; You can see the expected result ;
RFmx LTE/LTE-Advanced can measure and demodulate the LTE TDD signal success ;