Error -201366 When Running a Multidevice Task in NI-DAQmx
Updated Sep 27, 2024
  • NI-DAQ™mx
When I run an NI-DAQmx Analog Input task in LabVIEW or NI-MAX, using multiple identical modules in the same task (multidevice task), I get the following error message:
Error -201366 occurred at DAQmx Start Task.vi.
Possible reason(s):
Reference Clock signal was not found at the specified source.
This issue could indicate that there is a hardware problem in one of the modules used in the multidevice task, which does not allow the modules to be synchronized and run together as part of the same task.
To confirm there is an issue with any of the modules, follow these troubleshooting steps using a Test Panel in NI MAX , or Create a Task in NI MAX. This will help to isolate issue from the code:
1. Run your Analog Input task in each module separately to confirm each module is working fine individually.
2. Run the Analog input task including channels from different modules. If you get the error, repeat this step with different pairs of modules until you find which module is causing this issue in a multidevice task.
3. If you find one module that is causing this issue, change the module to a different slot to isolate this issue from the PXI slot where it is connected. If the issue follows the module, then the module needs to be sent for repair. If the issue follows the PXI slot, then the PXI chassis needs to be sent for repair.