What Is the Behavior of the USB Ports on a PXIe-8135 Controller in Case of a Short (By Connected Dut)?
Updated Feb 19, 2025
  • PXIe-8135
We are connecting a DUT to a USB-Port of the PXIe-8135 controller. How does the controller behave in case the DUT shorts the USB power supply lines (VBUS and GND)?
A USB power distribution switch is used in the USB design of the PXIe-8135. It implements multiple protection features one of which being Short-Circuit Protection. When the output voltage drops below 1.5V, which is caused by an over-load or a short-circuit, the devices limit the output current down to a safe level. The short-circuit current limit is used to reduce the power dissipation during short-circuit conditions. If the junction temperature reaches over-temperature threshold, the device will enter the thermal shutdown (Over-Temperature Protection).
We did a test during V&V on the limits and short conditions using E-load to verify the protection circuitry. The limits are the following:
Two side-by-side USB2.0 ports shared about 1.7A typical current limit (e-fuse limit range between 1.5A to 2.0A).
Two side-by-sideĀ USB3.0 ports shared about 2.3A typical current limitĀ (e-fuse limit range between 1.9A to 2.5A).
The limits are similar among all PXIe controllers.