Error -1074118647 When Performing PXI-4110 Calibration in Calibration Executive
Updated Jul 19, 2023
  • PXI-4110
  • Calibration Executive
When performing the the PXI-4110's calibration procedure on Calibration Executive I get the following error at the Verify Voltage Line Regulation Step:
Error -1074118647 occurred at niDCPower Configure Output Enabled.vi at step Verify Voltage Line Regulation. The auxiliary power source was disconnected
How can I solve this error?
This error appears if the auxiliary power supply of the PXI-4110 is not connected properly or if it is faulty.
In order to solve this issue you can perform the following actions:
  • Replace the auxiliary power supply of the PXI-4110.
  • Ensure the auxiliary power supply is able to produce the required voltage and current limits as stated on the calibration procedure for the device.
  • Reconnect the cables going into the auxiliary power supply connector of the PXI-4110.