Can I Take Thermocouple Measurements With Regular Analog Input Modules?
Updated Jul 13, 2026
  • NI-9211
  • NI-9212
  • NI-9213
  • NI-9214
  • PXIe-4353
  • I am trying to get some thermocouple measurements using my NI-9201, which is a regular analog input module. When I do this, I get erroneous/unexpected thermocouple measurements, how can I fix it?
  • Can I use NI-9220, which is an analog input module, for K-type thermocouple measurements?
Using regular analog input (AI) modules is not recommended for taking thermocouple measurements.
The resolution and voltage range for most of the AI modules are not suitable for thermocouple measurements. For example, the NI-9201 has fixed voltage range of +-10 V and a resolution of 12 bits. This means that the maximum resolution achievable is of 20/2^12 = 4.88 mV. Thermocouples vary in the order of uV/°C given by the Seebeck coefficient (i.e. 56uV/°C for a J-type).
Using temperature modules will give much better thermocouple measurements. See our sales page for our current temperature module offerings. You can also find temperature modules in the PXIe form factor such as the PXIe-4353.
Other factors such as accuracy and temperature drift must be considered in thermocouple measurements. These factors will deprecate even more your thermocouple measurements when using regular AI modules.