このコンテンツは設定された言語で表示できません。
Understanding Pseudodifferential Input Configuration and Supported NI Hardware
Updated Aug 31, 2026
  • Sound and Vibration Device
  • My device's specifications state that my hardware supports a pseudodifferential Terminal Configuration. What does it mean?
A pseudodifferential connection is a configuration that combines key advantages of differential and single-ended measurement systems.
Like a true differential input, a pseudodifferential setup uses two signal lines (a positive input terminal AI+ and a negative reference terminal AI-) to take a potential difference measurement. However, unlike true differential signals where both lines carry complementary signals, the negative terminal (AI-) in a pseudodifferential configuration does not carry a varying signal of interest—it serves solely as a DC reference point.
In pseudodifferential mode, the negative lead (AI-) is connected through a low-impedance path (often via a ground-referencing resistor or a 50Ω connection to chassis ground) to minimize ground loops while rejecting common-mode noise.
Dynamic Signal Acquisition (DSA) devices (such as the PXIe-446x, PXIe-448x, PXIe-449x series, and C Series NI 9230/9232/9233/9234) frequently utilize pseudodifferential input configurations due to the physical and electrical design of 2-wire Integrated Electronics Piezo-Electric (IEPE) sensors (e.g., accelerometers and microphones) and single-ended BNC/coaxial cabling connections.
The exact reasons for using pseudodifferential configurations with DSA cards include:
  1. IEPE Current Loop Return Path Requirement: 2-wire IEPE sensors require a constant current excitation (typically 2 mA to 4 mA) supplied on the positive terminal (AI+). To complete the circuit, the excitation current requires a defined return path to ground on the negative lead (AI-). In a true differential configuration, the negative lead floats, which interrupts the IEPE current return path and triggers driver errors (e.g., DAQmx Error -200077). Pseudodifferential mode provides an internal low-impedance path (typically a 50 Ω resistor to chassis ground) on AI- to complete the excitation current loop while allowing excitation to function properly.
  2. Ground Loop Mitigation: Case-grounded or field-grounded sensors often introduce subtle potential differences between the sensor ground and the measurement chassis ground. A standard single-ended configuration directly couples these ground potential differences into the signal path, creating ground loops that cause 50/60 Hz powerline hum or high-frequency switching noise. Pseudodifferential mode references AI- to ground through a 50 Ω path, which breaks direct low-impedance ground loops and suppresses common-mode ground noise.
  3. Compatibility with Coaxial / BNC Cabling: Most DSA applications rely on coaxial BNC or 10-32 connectors. The center pin connects to AI+ (signal and current excitation) and the outer shell connects to AI- (signal reference and current return). Pseudodifferential mode allows single-ended coaxial cabling to benefit from common-mode noise rejection without requiring fully balanced 3-wire differential cabling.
