If less current draw from the DUT is required when it's not actively being tested, there are a few options to eliminate or minimize the current draw.
For the PXI-4132, PXIe-4154, PXIe-4136/4137, and PXIe-4138/4139 only, you can set the Source:Output Connected property node to False to disconnect the DUT from the SMU by opening the SMU's on-board relay.  This will make the SMU transparent to the DUT so the SMU can't sink any current.
For the PXIe-414x, PXIe-4136/4137, and PXIe-4138/9 only, you can set the current level to 0 A in Constant Current mode. This will make the SMU appear 'open' and no current will flow from the DUT into the SMU. Note: only the PXIe-414x, PXIe-4136/4137, and PXIe-4138/9 will allow you to set the current level to 0; all other NI SMUs have a minimum current level, typically 2% of the current range.
For the PXI-4130, to minimize current draw from the DUT, leave the output enabled in Constant Current mode and set the current range to 200 uA.  Setting the SMU to the smallest current range allows you to set the current level to its smallest possible value. For the PXI-4130, in the 200 uA range the minimum programmable current level is 4 uA.  If 4 uA is still too large of a current draw from your DUT, you may need to consider installing an external relay to disconnect your DUT from your PXI-4130.