Measuring Current with an Analog Voltage Input
Updated May 9, 2019
  • Multifunction I/O Device
  • DAQ (Legacy)
  • Current Shunt
How do I measure current with an analog voltage input?
Do I need a shunt resistor in my current measurement circuit?
Current cannot be measured directly using an analog input. However, if a known resistance (shunt resistor) is placed in series with the load and the voltage across it is measured, the current through the load can be calculated using Ohm's Law.
Figure 1: Ohm's Law
The measurement accuracy for this configuration is highly dependent on the resistor accuracy. Any resistors are acceptable, as long as the maximum expected current multiplied by the resistance does not exceed the input range of your input device.
When measuring current in this fashion, the smallest value of shunt resistor compared to the load resistance should be used. This will create the smallest interference with the existing circuit. However, smaller resistances create smaller voltage drops, so a compromise between resolution and circuit interference must be made.
Note: To avoid damaging the shunt resistor, make sure the current passing through it does not exceed the current specifications for the shunt resistor.
Figure 2: How to Connect a Shunt Resistor to a DAQ Board
Placement of the shunt resistor in the circuit is important. If the external circuit shares a common ground with the computer containing the analog input device, then the shunt resistor should be placed as close to the ground leg of the circuit as possible, as shown in figure three.
Figure 3: Correct Usage of a Shunt Resistor
If not, the common mode voltage produced by the shunt resistor may be outside of the specification for the analog input device. This can lead to inaccurate readings or damage to the board, as shown in Figure 4.
Figure 4: Incorrect Usage of a Shunt Resistor
For such applications, National Instruments provides the SCXI Resistor Kit (a set of four precision resistors at 249 Ω with ±0.1% accuracy) for current measurement applications.
Some Multifunction I/O devices have internal shunt resistors, which have a known resistance value. This can be selected upon configuration of a virtual channel, instead of setting up a circuit with an external resistor.