Error 1074118655, 1073807343 or 1074118644 When Addressing a PXI Device
Updated Apr 24, 2026
  • PXIe-5110
  • NI-DMM
  • NI-SCOPE
  • NI-FGEN
  • NI-DCPower
  • NI-SWITCH
  • NI-HSDIO
  • NI-VISA
  • IVI Compliance Package
When I try to open a session to my PXI instrument (NI-DMM, NI-FGEN, NI-SCOPE, NI-HSDIO, NI-SWITCH, NI-DCPower) or try to validate my virtual device in NI Switch Executive, I get error 0XBFFF0011 or one of the following errors:
DriverError Code
NI-DMM-1074118655
NI-DCPower-1073807343
NI-SWITCH-1074118644
NI Switch Executive-1074118644
NI-FGEN-1073807343
NI-SCOPE-1073807343
NI-HSDIO-1073807343
These errors can have multiple causes. Try the following suggested troubleshooting steps below:
  • These errors usually occur when you address the instrument with a VISA resource descriptor (e.g. PXI1::15::INSTR) instead of a DAQ or DAQmx resource descriptor (e.g. DAQ::1::INSTR for Traditional DAQ, Dev1 or PXI1Slot1 for DAQmx). Make sure that you are using the correct resource descriptor for the driver you are using to communicate with your device.
  • These errors can also occur when there is a mismatch between your device and driver (e.g. Opening a NI-FGEN session with a PXI DAQ card instead of using NI-DAQmx). Check to make sure you are using the appropriate driver for your hardware.
  • These errors can be the result of outdated drives. Check the Readme files for your drivers to verify system compatibility.
  • These error codes can also be caused by corrupted drivers. Consider formatting the PC.
  • These errors can be caused by a corrupted MAX database. This can happen for a number of reasons including sudden power loss in the system. Refer to the article Reset and Fix the Measurement & Automation Explorer Database File for instructions on how to reset a corrupt database.
  • If you are deploying the code to a Real-Time target, the controller might take some time to boot and establish the connection with the card. Add a 10-second delay at the beginning of your VI.
  • These error codes have also been encountered if one of the modules mentioned above resides in a PXI chassis that has not been identified correctly. Refer to the article PXI Chassis Appears in NI MAX As PXI System (Unidentified) for more information.
For NI Switch Executive, the resource descriptor is in the IVI Driver Session on the Hardware tab. Double-click on the Resource Descriptor to change it.
Note: The Resource Descriptor dropdown menu will only display the VISA resource descriptor; the DAQ/DAQmx resource descriptor will need to be manually entered.
Refer to the Hardware Asset section of the Measurement & Automation Explorer Help for valid resource descriptor options. The help can be accessed within Measurement & Automation Explorer (MAX) under Help >> MAX Help.