How To Use NI-DAQmx Switch Soft Front Panel of NI Resistor Module
Updated Jul 31, 2019
  • Switch Executive
I don't know much about the configuration process of the 2722 in the MAX soft front panel. What is the correspondence between Channels and Relays?
The PXI-2722 is a PXI Programmable Resistor Module. There are four kinds of resistor module: PXI-2720, PXI-2722, PXIe-2725, PXIe-2727. The product details can be found in PXI Programmable Resistor Module. If you want to configure the module, the NI-Switch driver software needs to be installed on the host. After the installation is complete, click the module to be configured in MAX. Click the Test Panel to open the NI-SWITCH soft panel which is shown in Figure 1 and 2.
Figure 1. NI-SWITCH Soft Front Panel
Figure 2. Soft Front Panel of PXI-2722
There are two tabs: the Channels tab and the Relays tab. In the Channels tab, you can set the connection relationship of two channels. The Relay tab will automatically set the state of the corresponding relay according to the current connection results.
Different modules include different number of banks. Take one bank (b0) as an example. The following table lists the channel connections if you want to control the relays related to b0.
x stands for different resistance or channel. Other bank connections are similar to the above table. This method is not only suitable for 2722, but also other resister modules.
Note that it is best to set Channel 1 to the upper bank selection when connecting, and Channel 2 is the specific relay selection to avoid configuration errors.