Why is My NI 9237 Calibration Failing?
Updated Jun 4, 2018
  • NI-9237
  • Calibration Executive
I am using Calibration Executive to calibrate an NI 9237 module, and the Verify Analog Input test is failing. Why is this?
If you have multiple channels on the 9237 module connected to test fixtures, this prevents the module from providing enough power to reach the proper excitation levels. Your test will then fail, because the test value and measured values will be off. Only connect one channel at a time to any test fixturing to properly run the calibration procedure.
The 9237's excitation limitations is described further in this article.