Replace Fluke 5700A in Calibration Procedure for Multifunction DAQ Devices
Updated Nov 8, 2018
  • Multifunction I/O Device
  • Calibration Executive
NI recommends in the B/E/M/S/X Series Calibration Procedure  to use Fluke 5700A as the calibrator. This instrument is end-of-life. What is the recommended replacement?
Fluke's recommended replacement for the 5700A is the 5730A.
The B/E/M/S/X Series Calibration Procedure  lists the requirements for the calibration equipment and instruments if the recommended models are not available. However, not all instruments are easily automated through NI Calibration Executive, so if using it to automate your calibration procedures, you should be prepared to use it in manual mode when using an unsupported instrument.