Maximum and Minimum Values for the Interchannel Delay Setting
Updated Aug 8, 2022
  • PCIe-6363
I would like to know what are the maximum and minimum values for the interchannel delay that I can set up for my DAQ Device.
Each DAQ device has different minimum and maximum settings. However, you can derive these values very easily.
Maximum interchannel delay = 1/(scan clock rate) divided by the number of channels in your scan list. For example, if you scan all 16 channels at a scan rate of 10,000 scans per second, your maximum interchannel delay would be (1/10,000)/16 = 6.25E-6 seconds.
Minimum interchannel delay = 1/(maximum sampling rate in samples per second for your device). For example, if you have a PCIe-6353 device, the maximum sampling rate is 1.25 MS/s. Therefore, the minimum interchannel delay would be 1/1.25E6 = 8.0E-7 seconds.